Quantitative PCR assays for sensitive and specific detection of Bundibugyo ebolavirus

Author List: Kyle McMahon1, Stella Nielsen1, Hannah Knoll1, Resham Talwar1, Davina Thompson1, Colby Wilkason1, Al Ozonoff1,2,3, Elyse Stachler1†, Pardis C. Sabeti1,4,5†

Author Affiliations:

1Broad Institute of MIT and Harvard, Cambridge, Massachusetts, USA

2Boston Children’s Hospital, Boston, Massachusetts, USA

3Harvard Medical School, Boston, Cambridge, Massachusetts, USA

4Harvard University, Cambridge, Cambridge, Massachusetts, USA

5Howard Hughes Medical Institute, Chevy Chase, Maryland, USA

†These authors have contributed equally to this work

Corresponding author: Elyse Stachler, estachle@broadinstitute.org

Summary

In May 2026, an outbreak of Bundibugyo ebolavirus (BDBV) in the Democratic Republic of the Congo was declared a Public Health Emergency of International Concern by the World Health Organization, underscoring the need for reliable, rapidly deployable diagnostic tools. Here, we present the development and analytical validation of reverse-transcriptase quantitative PCR (RT-qPCR) assays designed for sensitive and specific BDBV detection. We developed and evaluated three assay formats: a singleplex TaqMan assay, a duplex TaqMan assay incorporating a human internal control, and a singleplex SYBR assay to reduce dependence on probe availability during outbreak response. We evaluated assay performance using synthetic RNA gene fragments, whole viral RNA, and contrived clinical samples. Both TaqMan assays achieved a 95% limit of detection (LOD95) of 5 copies/reaction, while the SYBR Green assay achieved an LOD95 of 50 copies/reaction. In addition, the assays accurately detected BDBV viral RNA and did not detect Zaire ebolavirus or Sudan ebolavirus viral RNA under the conditions tested. By making the assay protocols and resources openly available through protocols.io, this assay set provides a flexible and accessible molecular detection tool for BDBV surveillance, research, and outbreak response.

Context

On May 5, 2026, the Ministry of Health of the Democratic Republic of the Congo (DRC) alerted the World Health Organization (WHO) to an outbreak of an unknown illness in Mongbwalu Health Zone, Ituri Province, following the deaths of four health workers within four days (WHO, 2026). On May 15, 2026, rapid response teams confirmed the outbreak as Ebolavirus Disease (EVD) caused by Bundibugyo virus (BDBV) (Democratic Republic of the Congo, Ministry of Public Health, 2026). Two days later the WHO declared the outbreak to be a Public Health Emergency of International Concern (PHEIC) due to the geographic spread in the DRC and Uganda, and a rapidly rising caseload (WHO, 2026). As of June 6, 2026, the DRC and Uganda Ministries of Health have reported 534 laboratory-confirmed cases and 93 confirmed deaths across the two countries (CDC, 2026). The rapid increase in suspected cases and fatalities has heightened public health concerns and underscored the need for reliable, rapidly deployable diagnostic tools.

Orthoebolavirus is a filovirus genus comprising six genetically distinct species of filamentous, linear, non-segmented negative sense, single-stranded RNA viruses (Cheng et al., 2025). Four species cause EVD in humans: Orthoebolavirus zairense, or Zaire ebolavirus (Z-EBOV); Orthoebolavirus sudanense, or Sudan ebolavirus (S-EBOV); Orthoebolavirus bundibugyoense, or Bundibugyo ebolavirus (BDBV); and Orthoebolavirus taiense, or Taï Forest ebolavirus (CI-EBOV) (CDC, 2026). The two remaining species circulate only in animal hosts: Orthoebolavirus restonense, or Reston ebolavirus (RESTV) and Orthoebolavirus bombaliense, or Bombali ebolavirus (BOMV) (CDC, 2026). The May 2026 outbreak is the 17th Ebola outbreak reported in the DRC since 1976 and the third outbreak globally attributed to BDBV. From 1976 through 2025, most reported EVD outbreaks, cases, and deaths were attributed to Z-EBOV, followed by S-EBOV, BDBV, and CI-EBOV (CDC, 2026).

Since Z-EBOV has caused the largest number of EVD outbreaks and has historically been associated with high case fatality rate, most commercially available, validated molecular diagnostics, vaccines, and therapeutics have focused on Z-EBOV, followed by S-EBOV and BDBV (Bettini et al., 2023 and WHO, 2024). This creates challenges during BDBV outbreaks, as assays designed for Z-EBOV and S-EBOV may not detect BDBV, allowing for continued early BDBV spread amongst communities (WHO, 2026). As patient cases continue to increase, significant gaps still remain for readily-available, robust diagnostics. Commercial reverse-transcription quantitative PCR (RT-qPCR) assays for BDBV detection are available, but open, non-proprietary options remain limited (Bettini et al., 2023). This creates challenges during a large and rapidly evolving outbreak, when proprietary designs may be difficult to independently assess against newly generated outbreak sequences and high demand may create supply chain constraints. Additionally, since these assays are proprietary, it is unknown if primer or probe mismatches to circulating viruses can also limit diagnostic performance and increase the risk of detection gaps as the outbreak evolves.

The limited availability of open, non-proprietary BDBV assay options motivated us to design RT-qPCR assays that could be optimized on both available sequences from the circulating outbreak strain and historical BDBV genomes, with the goal of creating a robust, transparent, and readily deployable detection tool. We implemented the assay in complementary formats suited to different operational settings: a high-sensitivity probe-based singleplex assay, a duplex assay that pairs viral detection with an internal human control, and a dye-based alternative that can be deployed when probe synthesis or supply is a limiting factor. We evaluated these formats using materials and workflows that can be rapidly reproduced by other laboratories, including synthetic RNA gene fragments, quantified standards, viral RNA, and contrived clinical samples generated in a human plasma background. We report the design process, analytical evaluation, and implementation protocols for these assays to support near-term outbreak response and future BDBV surveillance and research.

Results

In silico sensitivity and specificity analysis

We designed RT-qPCR primers and probes (sequences provided in Table 1) to detect BDBV and paired the viral assay with a human internal control assay adapted from Meddeb et al., 2019. The BDBV assay targets a 103 bp segment of the large polymerase (L) gene, which encodes the viral RNA-dependent RNA polymerase. At the time of design, the BDBV primers and probes have demonstrated 100% sequence identity to all 26 complete BDBV genomes available from NCBI Virus (Figure 1), as well as to 16 BDBV genome available through Pathoplexus as of June 4, 2026. In silico specificity analysis predicted no cross-reactivity of the BDBV primers and probes to Z-EBOV and S-EBOV genomes.

Table 1: Primer and probe sequences for Bundibugyo ebolavirus (BDBV) and mitochondrial circular DNA (mcirDNA) RT-qPCR assays. Target region, primer and probe names, sequences, and source for the BDBV assay and the human mcirDNA internal control assay.

Figure 1: Alignment of the Bundibugyo ebolavirus (BDBV) RT-qPCR primers and probe to historic BVBV genomes. Complete historical BDBV genomes were aligned in Geneious; a representative view of the alignment is shown for readability. The locations of the forward primer, probe, and reverse primer are indicated at the top of the alignment. Although sequence variation is present at some positions in the surrounding genomic region, all three primer and probe binding sites show 100% sequence identity across the historical genomes. The same primer and probe sequences also showed 100% sequence identity to available current outbreak genomes (n=16) as of June 4, 2026 (data not shown).

Analytical performance of TaqMan qPCR validated assays

We evaluated the TaqMan RT-qPCR assays using synthetic RNA gene fragments at known concentrations, whole viral RNA, and contrived clinical samples, defined here as whole viral RNA spiked into a normalized healthy human plasma background. The singleplex BDBV TaqMan assay exhibited an amplification efficiency of 96.6%, while the duplex BDBV assay exhibited an amplification efficiency of 93.9% (Figure 2). Both assays achieved a 95% limit of detection (LOD95) of 5 copies/reaction (Table 2). The human mitochondrial circular DNA (mcirDNA) internal control assay also performed similarly in singleplex and duplex formats, with amplification efficiencies of 103.5% and 104.1%, respectively. All standard curves had R2 values ≥ 0.99.

The BDBV assay also showed linear detection of whole viral BDBV RNA across serial dilutions in contrived clinical samples in both singleplex and duplex TaqMan formats (Figure 3, R2 ≥ 0.99). In addition, the BDBV TaqMan assay detected BDBV whole viral RNA and did not detect whole viral RNA from Z-EBOV or S-EBOV under the conditions tested. In the duplex assay, the mcirDNA internal control produced a consistent cycle threshold (Ct) across BDBV dilutions, as expected from the constant human plasma background. For all TaqMan assays, we selected 400nM forward primer, 400nM reverse primer, and 200nM probe as the optimal primer-probe concentration after down-selection based on standard curve efficiency within a target range of 90%–110%, linearity, and observed sensitivity using synthetic RNA gene fragment material.

Figure 2: Standard curves for Bundibugyo ebolavirus (BDBV) and mitochondrial circular DNA (mcirDNA) TaqMan RT-qPCR assays in singleplex and duplex formats. We generated standard curves for a) BDBV and b) the mcirDNA internal control assay using synthetic RNA gene fragments across serial dilutions. Data points show the mean and standard deviation of triplicate reactions. Solid lines show simple linear regressions, and dotted lines show the 95% confidence interval for the regressions. E indicates the RT-qPCR standard curve efficiency.

Table 1: Limit of detection (LOD) analysis for Bundibugyo ebolavirus (BDBV) RT-qPCR assays. We evaluated the 95% limit of detection (LOD95) for the singleplex TaqMan, duplex TaqMan, and SYBR Green BDBV assays using synthetic RNA gene fragments quantified by digital PCR. We defined the LOD95 as the lowest concentration that produced amplification in at least 95% of replicate reactions (n=21).

Figure 3: Detection of Bundibugyo ebolavirus (BDBV) whole viral RNA in contrived clinical samples. We generated contrived clinical samples by spiking BDBV whole viral RNA into normalized healthy human plasma background at three concentrations prepared by 10-fold serial dilutions of stock RNA. The BDBV assay showed linear detection across the dilution series in both singleplex and duplex formats. For the duplex assay, the mitochondrial circular DNA (mcirDNA) internal control showed a consistent cycle threshold (Ct) across BDBV dilutions, reflecting the constant human plasma background. Data points show the mean and standard deviation of triplicate reactions. Solid lines show simple linear regressions, and dotted lines show the 95% confidence intervals for the regressions.

Analytical performance of the SYBR Green RT-qPCR assay

We also evaluated the BDBV assay in a singleplex SYBR Green RT-qPCR format to provide an alternate detection modality that does not require a TaqMan probe, which can become a time-limiting reagent during rapid response scenarios. The SYBR Green assay had an amplification efficiency of 99.6% (R2 ≥ 0.99) and achieved an LOD95 of 50 copies/reaction (Figure 4, Table 2). We selected 150nM forward and reverse primer concentrations as the final condition after down-selection from tested concentrations based on standard curve efficiency within the target range of 90%–110%, linearity, and observed sensitivity using synthetic RNA gene fragment material. The SYBR Green assay also detected BDBV whole viral RNA in contrived clinical samples and did not detect whole viral RNA from Z-EBOV or S-EBOV, indicating the assay is still specific even without the added sequence discrimination provided by a TaqMan probe.

Figure 4: Standard curve for the Bundibugyo ebolavirus (BDBV) singleplex SYBR Green RT-qPCR assay. We generated the standard curve using synthetic RNA gene fragments across serial dilutions. Data points show the mean and standard deviation of triplicate reactions. The solid line shows a simple linear regression, and dotted lines show the 95% confidence interval for the regression. E indicates the RT-qPCR standard curve efficiency.

Conclusion

We present optimized and analytically validated RT-qPCR assays that expand open, non-proprietary options for sensitive and specific detection of BDBV. By implementing the assay as a singleplex TaqMan assay, a duplex TaqMan assay with a human plasma internal control, and a singleplex SYBR Green assay, we provide complementary formats that laboratories can adapt to different testing needs, reagent constraints, and outbreak-response settings. Across these formats, the assays achieved low limits of detection, detected BDBV RNA in contrived clinical samples, and did not detect Z-EBOV or S-EBOV RNA under the conditions tested. We provide protocols and assay resources through protocols.io to support rapid implementation by other laboratories. Together, these assays provide flexible tools for BDBV surveillance, outbreak response, and future research while reducing dependence on limited commercial assay options during high-demand public health emergencies.

Methods

Molecular assay design

Bundibugyo virus (BDBV) genomes (n=26) were downloaded from NCBI Virus with filters applied (Tax ID: 3052458, Nucleotide Completeness: complete, access date: May 17, 2026). These genomes were aligned in Geneious using MAFFT, and several TaqMan qPCR primer and probe combinations were designed using Geneious combined with Primer3. Once outbreak specific sequences were published, genomes were obtained from Pathoplexis (n=16 as of June 4, 2026). The chosen BDBV design had 100% sequence alignment to all historical BDBV sequences as well as to the newly released outbreak sequences. In addition, a previously published human internal control assay targeting circulating human mitochondrial DNA in human blood (mcirDNA) was adapted to a TaqMan qPCR assay and previously optimized (unpublished work). This assay was included as a human internal control for plasma samples.

In silico analysis of designs

Primers and probes were mapped to on-target and off-target sequences in Geneious Prime to test specificity in silico. An assay was predicted to detect a sequence if both primers and probe mapped with ≤ 3 mismatches. All assays were predicted to detect 100% of their on-target sequences and 0% of their off-target sequences. In addition, all primer and probe sequences were evaluated for broader cross-reactivity in NCBI BLAST.

Samples and controls
A synthetic double stranded DNA gene fragment (Twist Bioscience) representing the region of the BDBV genome the assay targets (L segment, RNA-dependent RNA polymerase) were in vitro transcribed (IVT) and DNase treated using the HiScribe® T7 High Yield RNA Synthesis Kit (New England Biolabs, E2040L) following manufacturer’s instructions. Post-IVT and DNase treatment, transcribed RNA material underwent purification using RNAClean XP beads (Beckman Colter, A63987) following manufacturer’s instructions. Purified RNA material was then quantified using Qubit™ RNA High Sensitivity (HS) Kit (Invitrogen, Q32852) as recommended by the manufacturer. Based on calculated quantification, the synthetic RNA material was diluted and normalized to 1E8 copies/𝜇L aliquots.

Synthetic double stranded DNA gene fragment (Twist Bioscience) targeting Mitochondrial Circular DNA (Cytochrome C Oxidase Subunit III) was used as a positive control for experimentation. This material was quantified using Qubit™ DNA High Sensitivity (HS) Kit (Invitrogen, Q33231) as recommended by the manufacturer. Based on calculated quantification, the synthetic DNA material was then diluted and normalized to 1E8 copies/𝜇L aliquots.

For simulated contrived sample experimentation whole viral RNA (Z-EBOV and S-EBOV provided by NEIDL, Boston University; BDBV obtained from BEI) was spiked into pooled extracted human plasma sample matrix obtained from healthy, non-infected individuals (Innovative Research). Briefly, human plasma was extracted utilizing the Quick-DNA/RNA MagBead Extraction Kit (Zymo Research, R2131) on the KingFisher™ Flex Magnetic Particle Processor (ThermoFisher) with the 96 Deep-Well Head as recommended by the manufacturer. This material was then pooled to create a standard and consistent human background signal in subsequent RT-qPCR experimentation.

Singleplex TaqMan qPCR primer-probe optimization

All assay designs underwent initial evaluation as singleplex FAM TaqMan qPCR assays at three different primer-probe concentrations (primer concentrations/probe concentration): 200nM/200nM, 400nM/200nM, and 600nM/200nM Probe (Integrated DNA Technologies). Evaluation and performance of these assays was conducted using the Luna Probe One-Step RT-qPCR Kit (No ROX) (NEB, E3007E) as 10𝜇L reactions in triplicate following manufacturer’s recommendations (see protocols for reaction composition and cycling conditions). Each primer-probe concentration was evaluated utilizing synthetic gene fragment RNA or DNA on a standard curve ranging from 1E7 copies/reaction–1E1 copies/reaction. Assay performance was then evaluated based on qPCR standard curve efficiency (ranging from 90%–110%), linearity, and observed sensitivity of the assay based on imputed material concentration. All qPCR experiments were conducted on a QuantStudio 6 Flex (Applied Biosystems).

Duplex optimization and evaluation

After selecting optimal primer-probe concentrations for each singleplex FAM TaqMan probe assay, a duplex assay composed of FAM (BDBV) and HEX (mcirDNA) was evaluated in singleplex and duplex form simultaneously utilizing each unique synthetic gene fragment RNA or DNA on a standard curve ranging from 1E7 copies/reaction–1E1 copies/reaction. Evaluation and performance of these assays was conducted using the Luna Probe One-Step RT-qPCR Kit (No ROX) (NEB, E3007E) as 10𝜇L reactions in triplicate following manufacturer’s recommendations. Duplex assay performance was compared to singleplex assay performance to ensure consistent sensitivity, qPCR standard curve efficiency, linearity, and general performance in a more complex, higher order multiplexing assay composition. All qPCR experiments were conducted on a QuantStudio 6 Flex (Applied Biosystems).

SYBR Green singleplex optimization and evaluation

A singleplex SYBR Green BDBV assay was optimized utilizing the following primer concentrations (forward and reverse primers at equimolar concentrations): 100nM, 150nM, 200nM, and 450nM (Integrated DNA Technologies). Evaluation and performance of this assay was conducted using the Power SYBR™ Green RNA-to-CT™ 1-Step Kit (Applied Biosystems™, 4389986) as 10𝜇L reactions in triplicate following manufacturer’s recommendations (see protocol for reaction composition and cycling conditions). Each condition was evaluated utilizing synthetic gene fragment RNA on a standard curve ranging from 1E7 copies/reaction to 1E1 copies/reaction. Assay performance was then evaluated based on qPCR standard curve efficiency (ranging from 90%–110%), linearity, and observed sensitivity of the assay based on imputed material concentration. All qPCR experiments were conducted on a QuantStudio 6 Flex (Applied Biosystems).

Limit of Detection Determination

To determine the limit of detection (LOD) of our optimized assays, we ran replicates (n=21) of the following synthetic gene fragment concentrations for the BDBV singleplex TaqMan assay, duplex assay, and SYBR assay: 5E1, 1E1, and 5E0 copies/reaction. The gene fragments were accurately quantified and normalized on digital PCR (dPCR) using the QIAcuity OneStep Advanced Probe Kit (Qiagen, 250131) following the manufacturer’s recommendations. The LOD95 was defined as the concentration where at least 20/21 replicates amplified. All qPCR experiments were conducted on a QuantStudio 6 Flex (Applied Biosystems).

Acknowledgements

The following reagent was obtained through BEI Resources, NIAID, NIH:

  • RNA from Bundibugyo ebolavirus, Prototype Isolate #811250 (200706291 Uganda), NR-31812

We would like to thank the NEIDL at Boston University for providing RNA from Zaire ebolavirus and Sudan ebolavirus.

Funding Statement

This work is supported by the John D. and Catherine T. MacArthur Foundation, Flu Lab, and a cohort of generous donors through TED’s Audacious Project, including the ELMA Foundation, MacKenzie Scott, the Skoll Foundation, and Open Philanthropy.

Conflicts of Interest

P.C.S. holds several patents related to diagnostic technologies and is a co-founder and equity holder in Delve Biosciences and Lyra Labs, a board member and equity holder in Polaris Genomics, and an equity holder of NextGenJane. P.C.S was formerly a co-founder of Sherlock Biosciences and board member of Danaher Corporation, until December 2024. All potential conflicts are managed in accordance with institutional policy.

Supplemental Information

Table S1: Assay gene fragment sequences used as positive control sequences

Protocols (please note it may take a day or two for them to be public on protocol.io):

Appendix A: Protocol for Bundibugyo ebolavirus qPCR TaqMan singleplex assay

Appendix B: Protocol for Bundibugyo ebolavirus qPCR TaqMan duplex assay (Bundibugyo + human internal control)

Appendix C: Protocol for Bundibugyo ebolavirus qPCR SYBR singleplex assay

2 Likes

Analytical validation of a multiplex (four target) quantitative RT-PCR assay for sensitive and specific detection of ebolaviruses

Summary

Rapid molecular diagnostics are essential for both outbreak response and ongoing surveillance of ebolaviruses. In earlier work, we developed and analytically validated a targeted qPCR assay for Bundibugyo ebolavirus in response to the 2026 outbreak in the Democratic Republic of the Congo. Here, we extend that work by developing a four-target multiplex TaqMan RT-qPCR assay that differentiates Bundibugyo, Zaire, and Sudan ebolaviruses while incorporating a human internal control. The multiplex assay preserved analytical performance relative to singleplex formats, with amplification efficiencies of 95.8–100.5%, linear standard curves with R² values ≥ 0.99, and an LOD95 of 5 copies/reaction for all viral and internal control targets. In contrived clinical samples consisting of whole viral ebolavirus RNA spiked into healthy human plasma, the assay detected each on-target virus across serial dilutions and showed no unintended off-target detection among the ebolaviruses tested. These results provide a complementary diagnostic format for ebolavirus surveillance: whereas outbreak-specific assays can be rapidly updated and deployed once a causative virus is identified, multiplex assays can support broader screening, differential diagnosis, and preparedness before and between outbreaks.

Results

Analytical performance of a validated multiplex ebolavirus TaqMan RT-qPCR assay

We evaluated a four-target multiplex TaqMan assay designed to differentiate among three clinically important ebolaviruses while monitoring sample adequacy with a human internal control. The multiplex included assays targeting Zaire ebolavirus (Z-EBOV; FAM), Bundibugyo ebolavirus (BDBV; HEX), Sudan ebolavirus (S-EBOV; Cy5), and mitochondrial circular DNA (mcirDNA; ROX). Primer and probe sequences for all targets are provided in Table 3. For all assays, we selected 400nM forward primer, 400nM reverse primer, and 200nM probe as the optimal primer-probe concentration after down-selection based on standard curve efficiency within a target range of 90%–110%, linearity, and observed sensitivity using synthetic RNA gene fragment material.

We evaluated the analytical performance of the assays using synthetic DNA or RNA gene fragments at known concentrations, whole viral RNA, and contrived clinical samples, comprised of whole viral RNA spiked into normalized healthy human plasma. The multiplex assay exhibited the following amplification efficiencies: 100.5% (Z-EBOV, compared to singleplex efficiency of 99.1%), 98.4% (BDBV, compared to singleplex efficiency of 96.8%), 95.8% (S-EBOV, compared to singleplex efficiency of 98.8%), and 97.9% (mcirDNA, compared to 97.0%) (Figure 5). All standard curves had R2 values ≥ 0.99. In addition, all assays achieved a 95% limit of detection (LOD95) of 5 copies/reaction in the multiplex configuration, maintaining the same level of sensitivity as the assays in a singleplex format (Table 4).

The multiplex assay showed linear detection of whole viral BDBV, Z-EBOV, and S-EBOV RNA across serial dilutions in contrived clinical samples (Figure 6, R2 ≥ 0.98). As well, each assay specifically detected on-target whole viral RNA and did not detect off-target whole viral RNA of other ebolaviruses under the conditions tested. The mcirDNA internal control produced a consistent cycle threshold (Ct) across whole viral RNA dilutions, as expected from the constant human plasma background.

Table 3: Primer and probe sequences for Bundibugyo ebolavirus (BDBV), Zaire ebolavirus (Z-EBOV), Sudan ebolavirus (S-EBOV) and mitochondrial circular DNA (mcirDNA) used in the multiplex (4-plex) RT-qPCR assay. Target region, primer and probe names, sequences, and source for the BDBV, Z-EBOV, S-EBOV, and mcirDNA assays in the multiplex configuration. (Note: The Z-EBOV probe has locked nucleic acids (LNAs) as indicated by + followed by the base.)

Figure 5: Standard curves for Bundibugyo ebolavirus (BDBV), Zaire ebolavirus (Z-EBOV), Sudan ebolavirus (S-EBOV), and mitochondrial circular DNA (mcirDNA) TaqMan RT-qPCR assays in singleplex and multiplex (4-plex) formats. We generated standard curves for a) BDBV, b) Z-EBOV, c) S-EBOV, and d) the mcirDNA internal control assay using synthetic DNA/RNA gene fragments across serial dilutions. Data points show the mean and standard deviation of triplicate reactions. Solid lines show simple linear regressions, and dotted lines show the 95% confidence interval for the regressions. E indicates the RT-qPCR standard curve efficiency.

Table 4: Limit of detection (LOD) analysis for Bundibugyo ebolavirus (BDBV), Zaire ebolavirus (Z-EBOV), Sudan ebolavirus (S-EBOV), and mitochondrial circular DNA (mcirDNA) in the multiplex (4-plex) RT-qPCR assay. We evaluated the 95% limit of detection (LOD95) for the multiplex TaqMan BDBV, Z-EBOV, S-EBOV, and mcirDNA assays using synthetic DNA/RNA gene fragments quantified by digital PCR. We defined the LOD95 as the lowest concentration that produced amplification in at least 95% of replicate reactions (n=21).

Figure 6: Detection of Bundibugyo ebolavirus (BDBV), Zaire ebolavirus (Z-EBOV), and Sudan ebolavirus (S-EBOV) whole viral RNA in contrived clinical samples. We generated contrived clinical samples by spiking whole viral RNA of each ebolavirus species into healthy human plasma background at three concentrations prepared by 10-fold serial dilutions of stock RNA. Each assay showed linear detection across the dilution series and exhibited intended target detection with no unintended off-target detection to the other ebolavirus species tested. The mitochondrial circular DNA (mcirDNA) internal control showed a consistent cycle threshold (Ct) across each whole viral RNA dilution, reflecting the constant human plasma background. Data points show the mean and standard deviation of triplicate reactions. Solid lines show simple linear regressions, and dotted lines show the 95% confidence intervals for the regressions.

Methods

Molecular Assay Design

Previously designed Zaire ebolavirus (Z-EBOV) and Sudan ebolavirus (S-EBOV) assays were utilized to facilitate rapid multiplexing (ongoing work, unpublished). Briefly, Z-EBOV (n=135) and S-EBOV (n=121) genomes were downloaded from NCBI Virus (from collection date: 01/01/2022, access date: 02/06/2025, min length: 18000), aligned in Geneious Prime, and specific primers and TaqMan probes were designed to detect each species. Both assays are predicted to detect 100% of on-target sequences and 0% of off-target sequences by in silico analysis (see original post for testing parameters).

Samples and controls

Double stranded DNA gene fragments (Twist Bioscience) targeting specific, unique regions of Z-EBOV (Viral Protein 24) and S-EBOV (L segment, RNA-dependent RNA polymerase) were processed and used as reported above.

Multiplex optimization and evaluation

Singleplex primer-probe optimization, evaluation as a multiplex assay, and limit of detection (LOD) determination was conducted as reported in the initial post. The multiplex assay was run with the following fluorophore combinations: Z-EBOV (FAM), BDBV (HEX), S-EBOV (Cy5), and mcirDNA (ROX).

Table S2: Gene fragment sequences used as positive control sequences for the multiplex (4-plex) assay.

Appendix D: Protocol for ebolavirus qPCR TaqMan multiplex assay (Bundibugyo ebolavirus, Zaire ebolavirus, Sudan ebolavirus, and human internal control)

Summary

Rapid molecular diagnostics are essential for ebolavirus outbreak response but centralized qPCR testing can be difficult to sustain in mobile or resource-constrained settings. In earlier work, we developed and analytically validated a TaqMan RT-qPCR assay for Bundibugyo ebolavirus (BDBV) both with and without a duplexed internal control. Here, we extend that work by adapting these assays for near point-of-care (near POC) use. We validated the assays to run on a portable, small, and rapid qPCR machine (Mic qPCR cycler) using a streamlined “fast” cycling protocol that decreases the qPCR total runtime from 65 minutes to 35 minutes. In addition, we evaluated performance of the assays using a lyophilized qPCR mastermix which decreases cold chain dependence. Both the singleplex and duplex BDBV assays maintained analytical performance compared to the original validated conditions, with amplification efficiencies of 96.9-107.5%, linear standard curves with R² values ≥ 0.99, and an LOD95 of 5 copies/reaction for BDBV using the fast Mic cycling conditions and an LOD95 of 10 copies/reaction for BDBV using lyophilized mastermix. These results show that previously validated BDBV assays can be adapted to faster and more field-deployable workflows while preserving analytical sensitivity, supporting decentralized testing during future ebolavirus outbreaks.

Context

To support real-world adoption of Bundibugyo ebolavirus (BDBV) testing in decentralized settings, we adapted our established singleplex and duplex TaqMan RT-qPCR assays for near point-of-care (near POC) workflows. To do so, we validated the use of a more accessible mobile qPCR cycler with a streamlined protocol while maintaining assay performance and reducing the need for cold chain support via lyophilized reagents. Together, these adaptations address two major barriers to field testing during outbreaks: instrument portability and reagent stability.

We utilized the Mic qPCR cycler (Bio Molecular Systems), a four-channel rotor-based qPCR instrument with fast cycling capabilities. The Mic qPCR cycler has a small footprint (15cm x 15cm x 13cm) and can be powered by an external battery system, making it especially useful in mobile or resource constrained environments (Bio Molecular Systems, 2021). To minimize the need for cold chain support, we evaluated the shelf-stable lyophilized LyoPrime Luna® Probe One-Step RT-qPCR Mix. By combining a portable fast-cycling instrument with lyophilized reagents, we built a workflow that supports more rapid and accessible BDBV testing in decentralized outbreak-response settings.

Note: Figure and table numbering in this post continues from the preceding Bundibugyo ebolavirus assay development posts in this thread.

Results

Analytical performance of validated Bundibugyo ebolavirus TaqMan RT-qPCR assays on the Mic qPCR cycler

We evaluated whether our established singleplex (BDBV) and duplex (BDBV and mitochondrial circular DNA (mcirDNA)) assays could be adapted to the Mic qPCR cycler, a portable, fast-cycling, instrument suited for decentralized testing. To do this, we modified the existing RT-qPCR cycling conditions by reducing the hold times for reverse transcription, denaturation, and annealing (Table 5), while benchmarking assay performance against the previously validated QuantStudio 6 Flex workflow. The fast Mic protocol preserved equivalent assay performance and limit of detection (LOD), while significantly reducing total RT-qPCR run time from 65 minutes per run to 35 minutes per run.

Table 5: Modified fast thermocycling protocol for TaqMan RT-qPCR detection on the Mic qPCR cycler. We modified the existing RT-qPCR cycling conditions to produce a streamlined protocol compatible with the Mic qPCR cycler. By shortening the hold times for reverse transcription, denaturation, and annealing, we reduced the RT-qPCR run time from 65 minutes per run to 35 minutes per run while maintaining equivalent assay performance and limit of detection.

We evaluated the analytical performance of the assays on the Mic qPCR cycler using synthetic DNA or RNA gene fragments at known concentrations and contrived clinical samples, defined as whole viral RNA spiked into normalized healthy human plasma. Compared to performance on the QuantStudio 6 Flex, under the previously established RT-qPCR cycling conditions, the singleplex BDBV assay exhibited 105.6% amplification efficiency (compared to 96.6%) while the duplex assay exhibited 97.2% amplification efficiency for the BDBV assay and 97.1% amplification efficiency for the mcirDNA assay (compared to 93.9% and 101.1%, respectively) (Figure 7). All standard curves had R2 values ≥ 0.99. In addition, all assays achieved a 95% limit of detection (LOD95) of 5 copies/reaction in the singleplex and duplex configurations, matching the sensitivity observed for the corresponding QuantStudio 6 Flex workflows (Table 6).

The singleplex and duplex assays showed linear detection of whole viral BDBV RNA across serial dilutions in contrived clinical samples. In addition, neither assay detected whole viral RNA from Z-EBOV or S-EBOV under the conditions tested. For the duplex assay, the mcirDNA internal control produced a consistent cycle threshold (Ct) across BDBV dilutions, as expected from the constant human plasma background.

Figure 7: Standard curves for Bundibugyo ebolavirus (BDBV) and mitochondrial circular DNA (mcirDNA) TaqMan RT-qPCR assays in singleplex and duplex formats on the Mic qPCR cycler under fast cycling conditions. We generated standard curves for a) BDBV and b) the mcirDNA internal control using synthetic RNA or DNA gene fragments across serial dilutions with the modified Mic qPCR protocol. Data points show the mean and standard deviation of triplicate reactions. Solid lines show simple linear regressions, and dotted lines show the 95% confidence interval for the regressions. E indicates the RT-qPCR standard curve efficiency.

Table 6: Limit of detection (LOD) analysis for Bundibugyo ebolavirus (BDBV) and mitochondrial circular DNA (mcirDNA) in singleplex and duplex RT-qPCR assay formats on the Mic qPCR cycler under fast cycling conditions. We evaluated the 95% limit of detection (LOD95) for the singleplex and duplex TaqMan BDBV and mcirDNA assays using synthetic RNA or DNA gene fragments quantified by digital PCR. We defined the LOD95 as the lowest concentration that produced amplification in at least 95% of replicate reactions (n=21).

Analytical performance of validated Bundibugyo ebolavirus TaqMan RT-qPCR assays using lyophilized mastermix

We evaluated a lyophilized RT-qPCR mastermix (LyoPrime Luna® Probe One-Step RT-qPCR Mix) on both the QuantStudio 6 Flex and Mic qPCR cycler to reduce cold chain requirements for BDBV testing in decentralized settings. We tested both the singleplex (BDBV) and duplex (BDBV and mcirDNA) assay formats and benchmarked performance against the previously reported liquid RT-qPCR mastermix product (Luna Probe One-Step RT-qPCR Kit). The lyophilized mastermix maintained assay performance and linearity, and produced low LODs, reducing the need for cold chain support.

We evaluated the analytical performance of the assays using the lyophilized mastermix with synthetic DNA or RNA gene fragments at known concentrations. The singleplex BDBV assay exhibited 107.5% (QuantStudio 6 Flex) and 104.5% (Mic qPCR Cycler) amplification efficiency, while the singleplex mcirDNA assay exhibited 101.3% (QuantStudio 6 Flex) and 104.1% (Mic qPCR Cycler) amplification efficiency (Figure 8). The duplex assay exhibited 99.9% and 99.7% amplification efficiency for the BDBV and mcirDNA assays, respectively (QuantStudio 6 Flex), and 97.1% and 96.9% amplification efficiency for the BDBV and mcirDNA assays, respectively (Mic qPCR Cycler) (Figure 8). All standard curves had R2 values ≥ 0.99. In addition, mcirDNA maintained an LOD95 of 5 copies/reaction, while the BDBV assay exhibited an LOD95 of 5 copies/reaction using the lyophilized mastermix on the QuantStudio 6 Flex and and an LOD95 of 10 copies/reaction on the Mic qPCR cycler (Table 7).

Figure 8: Standard curves for Bundibugyo ebolavirus (BDBV) and mitochondrial circular DNA (mcirDNA) TaqMan RT-qPCR assays in singleplex and duplex formats using lyophilized RT-qPCR mastermix on the QuantStudio 6 Flex (A-B) and Mic qPCR cycler (C-D). We generated standard curves for a) BDBV on QuantStudio 6 Flex, b) the mcirDNA internal control on QuantStudio 6 Flex, c) BDBV on Mic qPCR cycler, and d) mcirDNA on the Mic qPCR cycler using synthetic RNA gene fragments across serial dilutions with lyophilized RT-qPCR mastermix. Data points show the mean and standard deviation of triplicate reactions. Solid lines show simple linear regressions, and dotted lines show the 95% confidence interval for the regressions. E indicates the RT-qPCR standard curve efficiency.

Table 7: Limit of detection (LOD) analysis for Bundibugyo ebolavirus (BDBV) and mitochondrial circular DNA (mcirDNA) in singleplex and duplex RT-qPCR assay formats using lyophilized RT-qPCR mastermix on the QuantStudio 6 Flex and the Mic qPCR cycler. We evaluated the 95% limit of detection (LOD95) for the singleplex and duplex TaqMan BDBV and mcirDNA assays using synthetic RNA or DNA gene fragments quantified by digital PCR and amplified with lyophilized RT-qPCR mastermix on the QuantStudio 6 Flex and the Mic qPCR cycler. We defined the LOD95 as the lowest concentration that produced amplification in at least 95% of replicate reactions (n=20).

Conclusion

Deployable, rapid, and reliable detection assays are critical for effective outbreak response. We adapted our previously validated Bundibugyo ebolavirus TaqMan RT-qPCR assays for near point-of-care compatible testing workflows. These adapted workflows maintained the analytical performance, sensitivity, and specificity of a traditional qPCR while reducing run time to approximately 35 minutes, using cold chain independent reagents, and running on a portable qPCR instrument. Together, these features could improve access to rapid and reliable BDBV testing in resource-constrained environments and provide an advantageous option for future outbreak response.

Methods

Mic qPCR cycler thermocycling optimization

We adapted our BVBD assays to run on the Mic qPCR cycler (Bio Molecular Systems), a portable, fast cycling compatible qPCR machine. We initially ran the thermocycling profile of the assay as originally run on the QuantStudio 6 Flex in our previous work, and systematically changed the cycling conditions to assess performance with the goal of shortening the overall run time. The conditions tested were 1) original profile, 2) shortening the denaturation hold time to 0 seconds, 3) shortening the annealing hold time to 10 seconds, and 4) shortening the reverse transcription step to 5 min and the annealing hold time to 5 seconds. After optimization, we ran all remaining experiments with the run profile of (4).

Analytical validation of Mic qPCR BVBD assays

We adapted our singleplex and duplex TaqMan assays to be run on the Mic qPCR cycler and validated them using our fast cycling conditions. All assays were evaluated in singleplex and duplex form utilizing synthetic gene fragment RNA or DNA on a standard curve ranging from 1E7 copies/reaction–1E1 copies/reaction, ensuring that standard curves maintained an efficiency between 90-110%. All experiments were completed with the Luna Probe One-Step RT-qPCR Kit (No ROX) (NEB, E3007E) as 10𝜇L reactions in triplicate Mic tubes.

Evaluation of a lyophilized RT-qPCR mastermix

We evaluated our BDBV TaqMan singleplex and duplex assays on an available lyophilized RT-qPCR mastermix (LyoPrime Luna® Probe One-Step RT-qPCR Mix with UDG, NEB, L4001P) to ensure compatibility. Each well of the lyophilized mastermix was rehydrated with 18𝜇L of water containing primers and probes to give final assay concentrations of 400nM primer and 200nM probe in a 20𝜇L reaction. Next, 2𝜇L of sample was added to each well and duplicate 10𝜇L reactions were run on the Mic qPCR cycler or QuantStudio 6 Flex

Limit of Detection Determination

To determine the limit of detection (LOD) of our assays, we ran replicates (n=21 for non-lyophilized mastermix or n=20 for lyophilized mastermix) of the following synthetic gene fragment concentrations for the BDBV TaqMan singleplex and duplex: 1E1 and 5E0 copies/reaction. The gene fragments were accurately quantified and normalized on digital PCR (dPCR) using the QIAcuity OneStep Advanced Probe Kit (Qiagen, 250131) following the manufacturer’s recommendations. The LOD95 was defined as the concentration where at least 20/21 (for non-lyophilized) or 19/20 (for lyophilized mastermix) replicates amplified.